    Re: Wanna see some cool copies?

    Quote Originally Posted by JVergin View Post
    It's just a chart with color bars. The only special thing about it is on one side of the test chart, it has the exact values of each color in case you needed to manually enter each one. It even comes in a shiny cover sandwiched in between two cardboard pieces.
    What's so special about it? Well it's the only way to calibrate the CCD. That's about it. =^..^=
    If you'd like a serious answer to your request:
    1) demonstrate that you've read the manual
    2) demonstrate that you made some attempt to fix it.
    3) if you're going to ask about jams include the jam code.
    4) if you're going to ask about an error code include the error code.
    5) You are the person onsite. Only you can make observations.

    blackcat: Master Of The Obvious =^..^=

  2. #12
    Not a service manager 2,500+ Posts Iowatech's Avatar
    Join Date
    Dec 2009
    Location
    Iowa
    Posts
    3,933
    Rep Power
    97

    Re: Wanna see some cool copies?

    Quote Originally Posted by blackcat4866 View Post
    It's fixed. It was the ISC PWB. My guess is that when the CCD registers the RGB data, it references some sort of color table. One table for the contact glass, and another table for the DP. Whatever form that data is stored as must somehow be corrupt, at least on the document processor side.

    The new ISC improved the image immediately, but the directional lines in the gray were still invisible until I ran the U411 calibrations.

    And it's a Kyocera, not Konica.

    And yes, it's definitely an image processing issue. =^..^=
    OOPS! I should have known that. Sorry.

  3. #13
    Senior Tech 250+ Posts
    Join Date
    Dec 2013
    Posts
    368
    Rep Power
    26

    Re: Wanna see some cool copies?

    I must admit i never had something like this on any Kyocera color machine. Thanks for sharing blackcat.

  4. #14
    Senior Tech 100+ Posts
    Join Date
    Jul 2014
    Posts
    188
    Rep Power
    22

    Re: Wanna see some cool copies?

    Quote Originally Posted by blackcat4866 View Post
    @wrangsdale: That's really interesting! What sort of software did you use to alter the the image color?
    I used GIMP specifically for that image since it's what I have available at work. It's one of the better freeware image manipulation software out there.

    Glad to hear that you solved it!

  5. #15
    Senior Tech 250+ Posts
    Wanna see some cool copies?


    Join Date
    Mar 2009
    Posts
    408
    Rep Power
    37

    Re: Wanna see some cool copies?

    We have the exact same problem on a 8353ci. Engine board was swapped I think then it caused this.

    The dp colours are fine duplex but anything that uses the platen glass has the weird purple tint to everything.

    The board you are talking about is the one directly connected to the lens under the glass?

  6. #16
    Master Of The Obvious 10,000+ Posts
    Wanna see some cool copies?

    blackcat4866's Avatar
    Join Date
    Jul 2007
    Location
    Lapeer, Michigan
    Posts
    22,471
    Rep Power
    463

    Re: Wanna see some cool copies?

    Quote Originally Posted by fishleg View Post
    We have the exact same problem on a 8353ci. Engine board was swapped I think then it caused this.

    The dp colours are fine duplex but anything that uses the platen glass has the weird purple tint to everything.

    The board you are talking about is the one directly connected to the lens under the glass?
    No sir, the ISC is on the back frame of the scanner (at least on the Alphard2 series).

    I don't have a service manual For the Iris2, but I suspect that the CCD is wired directly to the Engine PWB. All the image processing must happen there. =^..^=
    If you'd like a serious answer to your request:
    1) demonstrate that you've read the manual
    2) demonstrate that you made some attempt to fix it.
    3) if you're going to ask about jams include the jam code.
    4) if you're going to ask about an error code include the error code.
    5) You are the person onsite. Only you can make observations.

    blackcat: Master Of The Obvious =^..^=

  7. #17
    Master Of The Obvious 10,000+ Posts
    Wanna see some cool copies?

    blackcat4866's Avatar
    Join Date
    Jul 2007
    Location
    Lapeer, Michigan
    Posts
    22,471
    Rep Power
    463

    Re: Wanna see some cool copies?

    Quote Originally Posted by blackcat4866 View Post
    No sir, the ISC is on the back frame of the scanner (at least on the Alphard2 series).

    I don't have a service manual For the Iris2, but I suspect that the CCD is wired directly to the Engine PWB. All the image processing must happen there. =^..^=
    I suspect that you may not have purchased the CCD calibration sheets, or run the calibration. There is a data table that must be populated to define the CCD colors. =^..^=
    If you'd like a serious answer to your request:
    1) demonstrate that you've read the manual
    2) demonstrate that you made some attempt to fix it.
    3) if you're going to ask about jams include the jam code.
    4) if you're going to ask about an error code include the error code.
    5) You are the person onsite. Only you can make observations.

    blackcat: Master Of The Obvious =^..^=

  8. #18
    Senior Tech 250+ Posts
    Wanna see some cool copies?


    Join Date
    Mar 2009
    Posts
    408
    Rep Power
    37

    Re: Wanna see some cool copies?

    IMG_22102021_192503_(3000_x_2250_pixel).jpg
    signal-2021-12-08-21-59-13-514.jpg

    That's what it would print and then a photocopy of it off the glass.

    I don't think it has a connection board on the 8353ci. We changed the engine board and it returned to normal.

  9. #19
    Senior Tech 250+ Posts
    Join Date
    Apr 2017
    Location
    Manitoba
    Posts
    348
    Rep Power
    20

    Re: Wanna see some cool copies?

    Quote Originally Posted by fishleg View Post

    I don't think it has a connection board on the 8353ci. We changed the engine board and it returned to normal.
    The ribbon cable that goes from the image scanner carriage to the engine may have deformed contact pads that were making a poor contact or short. I have seen several instances of various ribbons over the years that get wonky with removal/reinstallation and have learned to inspect or re-trim on each handling of the connection. Most notably was a screen ribbon on an M3540idn new out of the box that had a purple screen background.