+ Reply to Thread
Results 1 to 3 of 3

Thread: Toshiba e-studio 650

  1. #1
    Technician makaule is on a distinguished road
    Join Date
    Jan 2009
    Posts
    46
    Rep Power
    3

    Toshiba e-studio 650

    Service recomended for SPC. Changed almost all parts (drum, developer and etc..) 20K ago. What can be wrong ?

  2. #2
    Trusted Tech 50+ Posts atudorica is on a distinguished road
    Join Date
    Oct 2007
    Location
    burnaby
    Posts
    79
    Rep Power
    5
    be sure you reset drum counter in pm mode
    main charge & grid are a factor
    some times is the sensor and it is expensive

  3. #3
    Service Manager 250+ Posts ivovb is building a good name for himself
    Join Date
    Apr 2008
    Location
    Sofia
    Posts
    278
    Rep Power
    6
    Quote Originally Posted by makaule View Post
    Service recomended for SPC. Changed almost all parts (drum, developer and etc..) 20K ago. What can be wrong ?
    [Corrective action when “ Service Recommended for SPC” is blinked]
    Check the control status of the surface potential sensor (05-242).
    <When “4” is displayed: sensor error>
    The sensor detection value is abnormal or the sensor output value is not changed even though
    the main charger bias value is changed.
    Is the connector of the surface potential sensor connected properly?
    Reconnect the connector and perform the “surface potential sensor control check”
    (described later).
    Is the main charger grid/wire attached poorly?
    Is the main charger attached poorly?
    Are leaks and such occurring?
    After removing, if any, dust and correcting the defect, perform the “surface
    potential sensor control check” (described later).
    Is the main charger grid bias output when the main charger grid bias output adjustment (05-
    210) is performed?
    Or, is the bias value is changed when the adjustment value is changed?
    1. Replace the high-voltage transformer and perform the “surface potential
    sensor control check” (described later).
    2. Replace the LCG board and perform the “surface potential sensor control
    check” (described later).
    Does the output value of the main charger grid bias consist with the setting value (within the
    acceptable range)?
    Readjust the main charger grid bias output adjustment (05-210) and perform the
    “surface potential sensor control check” (described later).
    Replace the surface potential sensor and perform the “surface potential sensor control check”
    (described later).
    Replace the LGC board and perform the “surface potential sensor control check” (described later).

    <<Procedure of the “surface potential sensor control check”>>
    Set the value for the number of detected abnormalities of the drum surface potential control (08-
    420) to “0”.
    Set the value for the drum surface potential setting (08-421) to “0” (valid).
    Set the value for the auto-start print volume setting 1 of the image quality control (08-803) to “0” (valid).
    Perform a test printing.
    Check the control status of the surface potential sensor (05-242) and number of detected
    abnormalities of the surface potential sensor (08-420).
    If the value for 05-242 is except “4” and the value for 08-420 is “0”, the status is normal.
    Return the value for the auto-start print volume setting 1 of the image quality control (08-803) to “2”.
    The check is completed.

    BR

+ Reply to Thread

Tags for this Thread

Bookmarks

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts