costel fasola
11-29-2014, 03:23 PM
at start sc 21-07
i try dipsw- 3-3 -1
3-4- 1
and 6-2-1
looks like "ready to copy" but if i try to cpoy is sc 95-01

if i try to do hv ajunstment : error adjustment
any ideea?

Luther
11-29-2014, 05:26 PM
Fan
abnormality
SC95-01 The EM signal is found abnormal a
specified period of time after the
ADU cooling fans /1 (FM14), /2
(FM15) and /3 (FM13), the sensor
cooling fan (FM16) and the registration
cooling fan (FM17) turn on.
And the signal is still found abnormal
a specified period of time after
it turns on again after turning off,
and the door close of the door
open/close sensors /1 (PS1) and /
2 (PS2) and the PF door open/
close sensor (PS70) has been
detected 10 times in a row.
ADU drive board /1
(ADJDB1)
ADU cooling fan /1 (FM14), /
2 (FM15), /3 (FM13)
Sensor cooling fan (FM16)
Registration cooling fan
(FM17)

21-07 is the transfer cleaning motor.either the cleaner is not moving or the motor is bad

laitang
11-30-2014, 02:11 AM
27280

costel fasola
11-30-2014, 07:59 AM
Fan
abnormality
SC95-01 The EM signal is found abnormal a
specified period of time after the
ADU cooling fans /1 (FM14), /2
(FM15) and /3 (FM13), the sensor
cooling fan (FM16) and the registration
cooling fan (FM17) turn on.
And the signal is still found abnormal
a specified period of time after
it turns on again after turning off,
and the door close of the door
open/close sensors /1 (PS1) and /
2 (PS2) and the PF door open/
close sensor (PS70) has been
detected 10 times in a row.
ADU drive board /1
(ADJDB1)
ADU cooling fan /1 (FM14), /
2 (FM15), /3 (FM13)
Sensor cooling fan (FM16)
Registration cooling fan
(FM17)

21-07 is the transfer cleaning motor.either the cleaner is not moving or the motor is bad
Still not work. i try change aduboard, hv board on adu , cleaning motor, transferasist motor. Same thing.
I try dpsw 3-3-1 3-4-1 and 6-2-1. Machine enter in "ready to copy" and in first page on print or copy go's in sc 95-01.
If i try to do hv ajustement ( with sw 3-3-1 3-4-1 6-2-1 ) : transfer error.
in state confirmation cleaning motor dont work with good motor from a god machine

costel fasola
12-01-2014, 07:15 PM
back to the issue. I changed to another copier Adu Adu unit. f is running fine on another machine. I guess the problem is something behind the copier can PRCB. got off PRCB I checked the ICP sites and have not found anything burned. I put it back and now surprise. A new code 93-01. all Adu something to the unit.
advice do you have?