    ricoh 2075 under toning

    Hi i'm new to the forum, could anyone help me. I have a Ricoh 2075 that keeps under toning after approx 1000 copies. I force toner supply and reset TDC and it always under tones again.

    Thanks in advance
    Papadave
  • cdog530
    Technician

    50+ Posts
    • Nov 2009
    • 66

    #2
    Re: ricoh 2075 under toning

    When was the developer last replaced?

    Comment

    • papadave

      #3
      Re: ricoh 2075 under toning

      Originally posted by cdog530
      When was the developer last replaced?
      hi cdog530 the dev was replaced 20 k ago. it lasted only 2 k & since then i have been toning up all the time.

      regards

      papadave

      Comment

      • TonerMunkeh
        Professional Moron

        2,500+ Posts
        • Apr 2008
        • 3865

        #4
        Re: ricoh 2075 under toning

        If you cheat the front door does the bottle actually turn as it's trying to add toner? Can you hear the toner add clutch click as it tries to move the auger in the sub hopper? Are you getting good readings from the sub hopper piezo sensor? Could be the toner add clutch has failed in the developer unit drive pack. Only down side is, you can't get the clutch on it's own, it's a complete assembly.
        It's 106 miles to Chicago. We've got a full tank of gas, half a pack of cigarettes, it's dark and we're wearing sunglasses.

        Hit it.

        Comment

        • banginbishop
          grumpy old git

          500+ Posts
          • Oct 2007
          • 894

          #5
          Re: ricoh 2075 under toning

          can you clarify what you mean by reset tdc? because you only run that once when dev is replaced.
          Incontinentia Buttocks

          Comment

          • papadave

            #6
            Re: ricoh 2075 under toning

            Originally posted by banginbishop
            can you clarify what you mean by reset tdc? because you only run that once when dev is replaced.
            Once i have toned the dev back to a reasonable black density, surely to try and retain that density you must give the machine a voltage reference. All I can think of doing is to carry out initial toner sensor density sp 2801

            Comment

            • banginbishop
              grumpy old git

              500+ Posts
              • Oct 2007
              • 894

              #7
              Re: ricoh 2075 under toning

              Ok for a start new bag of dev. You only run that once and when it's a fresh bag of dev. Then run Id settings in 3 something can't remember what off hand then print off an smc 5990 1 all then non default to check nothing's been altered.
              Incontinentia Buttocks