    Konica C364e scanner green stripes

    Hi everyone

    A client of mine reports that in the morning the first scans come out with green lines then after about 10 scan works the problem disappears.
    I thought of a CCD problem or of a connection cable do you have other possible solutions?

    defect in the attachment Green squares are only for privacy

    PS: the printer is in the office so not in an too cold environment.
    Attached Files
  • Synthohol
    Certified Konica Expert

    Site Contributor
    5,000+ Posts
    • Mar 2016
    • 5533

    #2
    are scans off the glass different?
    We know a thing or two because we've seen a thing or two.
    The medication helps though...

    Comment

    • ferro10
      Trusted Tech

      100+ Posts
      • Jun 2015
      • 109

      #3
      Originally posted by Synthohol
      are scans off the glass different?
      they are different as documents
      Or do you mean something else?

      Comment

      • Synthohol
        Certified Konica Expert

        Site Contributor
        5,000+ Posts
        • Mar 2016
        • 5533

        #4
        are you scanning through the feeder or off the platen glass?
        We know a thing or two because we've seen a thing or two.
        The medication helps though...

        Comment

        • ferro10
          Trusted Tech

          100+ Posts
          • Jun 2015
          • 109

          #5
          Originally posted by Synthohol
          are you scanning through the feeder or off the platen glass?
          That test is done with feeder. The customer confirms me that the glass scan also comes with the same defect.

          Comment

          • Synthohol
            Certified Konica Expert

            Site Contributor
            5,000+ Posts
            • Mar 2016
            • 5533

            #6
            ok so likely ccd related then or mirrors could be slightly out of place.
            you could reseat all the cables, ribbons on both ends but dont touch red screws.
            just for giggles, run background data reset and stabilize, then stabilize in gradations and perform all the adjustments (3 each) and see if the result changes.
            that test wont require a screwdriver
            We know a thing or two because we've seen a thing or two.
            The medication helps though...

            Comment

            • Woxner
              Senior Tech

              500+ Posts
              • Jul 2011
              • 815

              #7
              if it clears up just replace ccd. have had many do this. keep it simple. it cant be mirrors so dont piss with them

              Comment

              • myster
                Technician

                50+ Posts
                • Oct 2011
                • 79

                #8

                Good morning.
                We had a similar problem on a c652. In the morning the "first hour" failed, the rest of the day it worked fine.
                If it was turned off at night the next morning it would happen again
                We solved it by changing the ccd unit


                Comment

                • blackcat4866
                  Master Of The Obvious

                  Site Contributor
                  10,000+ Posts
                  • Jul 2007
                  • 22744

                  #9
                  On machines with high scan volume and very large packets, it's common for CCD failures and intermittent codes C9401, across multiple machine series'.
                  If you'd like a serious answer to your request:
                  1) demonstrate that you've read the manual
                  2) demonstrate that you made some attempt to fix it.
                  3) if you're going to ask about jams include the jam code.
                  4) if you're going to ask about an error code include the error code.
                  5) You are the person onsite. Only you can make observations.

                  blackcat: Master Of The Obvious =^..^=

                  Comment

                  • ferro10
                    Trusted Tech

                    100+ Posts
                    • Jun 2015
                    • 109

                    #10
                    Update

                    I confirm that the problem has been resolved by replacing the CCD